General => General Discussion => Topic started by: USMCPMEL on 11-15-2009 -- 14:32:18

Title: 34980A Calibration
Post by: USMCPMEL on 11-15-2009 -- 14:32:18
Just got 2 of these beasts in. Has anyone calibrated these before? Does it meet 4 to 1 with the 5520A?
Title: Re: 34980A Calibration
Post by: RFCAL on 11-17-2009 -- 15:10:40
We use a 5720A.I don't think the 5520 will suffice.
Title: Re: 34980A Calibration
Post by: Hawaii596 on 11-18-2009 -- 09:24:18
That sounds right.  We had some of them at my previous employer.  They are pretty accurate.  We had piles of 34970A's and they were in process of replacing them with 34980A's.
Title: Re: 34980A Calibration
Post by: HalAC on 11-18-2009 -- 17:51:17
a 5720 will be sufficient for the DC volts, current and ohms, but you'll need a 5790A to do the AC volt ranges along with a shunt set to do the AC current ranges. That is if you want to maintain a 4:1.
Title: Re: 34980A Calibration
Post by: USMCPMEL on 11-18-2009 -- 17:58:24
Dang I was hoping that was not the case. They looked kind of el cheapo I thought maybe they were not that accurate.
Title: Re: 34980A Calibration
Post by: CalLabSolutions on 11-18-2009 -- 21:52:41
You may be able to do it with a 5520 and a 3458A.  You will have to measure the output of the 5520 with the 3458A then compair those measurements with the 34980A.  It will be labor intensive, but it may be accurate enough to test the units with the standards you have.