Title: 8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE
Post by: wijogr on 06-02-2017 -- 12:01:23
I'm seeing a difference between using the 4-wire 200 Ohm LoI and 2 kOhm readings vs. the PRT resistance displayed on the right screen of the 8508A in PRT mode. I think I'm missing a setting but can't figure out what it is. The PRT measurement does not use FAST so I have that turned off in the resistance function. Ohms Guard is set for both functions. I=1 mA for all three. Low pass Filter is on. Both resistance ranges have been Zeroed. Any ideas?
Title: Re: 8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE
Post by: Mr M. on 06-03-2017 -- 18:48:36
Temperature uses 4-wire current reversal resistance measurement, this would be the equivalent of the TRUΩ setting.
Title: Re: 8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE
Post by: bAdbOb on 06-05-2017 -- 08:24:49
The current changes when you change ranges. I believe 1 mAdc is the usual current used for RTDs.
Title: Re: 8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE
Post by: NC-Cals on 06-05-2017 -- 11:04:08
The manual has you zero the 200Ω range in LOI (1 mA) and the 2kΩ range in 1 mA. This is outlined on page 3-27 of the user manual. You have to zero the ranges before you switch to PRT. There are also instructions on setting up for 3-wire PRT measurement.
Title: Re: 8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE
Post by: wijogr on 06-05-2017 -- 14:20:54
Thank you for all the suggestions. I was not using the TRU Ohms setting in resistance.  :-o After running two tests it is working perfectly. Thanks again.