Started by CalLabSolutions, 03-22-2013 -- 00:16:12

Previous topic - Next topic

CalLabSolutions

If you have a Fluke 9500 that is coming due calibration I recommend you add a 250k Ohm Calibration point to the resistance test points.. The newer scope procedures from Tektronix are adding a this test point to their verification procedures.

The 9500 does not spec this test point, so we have been working with our customers to test this point and establish the uncertainties.  What we have found is the 250k Ohm measurement uses the same range as the 1M Ohm measurement.  We have been successful testing this point to the same specifications as the 1M Ohm specs.

If you have any questions drop me a line.

Mike
Michael L. Schwartz
Automation Engineer
Cal Lab Solutions
  Web -  http://www.callabsolutions.com
Phone - 303.317.6670

Hawaii596

I wonder why the added point.  I cal my 9500's, and do a check at 50 Ohms and 1 Megohms, I thought because it is only used to verify input impedance at those two values. 
"I often say that when you can measure what you are speaking about, and express it in numbers, you know something about it; but when you cannot measure it, when you cannot express it in numbers, your knowledge is of a meagre and unsatisfactory kind."
Lord Kelvin (1824-1907)
from lecture to the Institute of Civil Engineers, 3 May 1883

RFCAL

This does not apply to the 9500B

Hawaii596

Mine are the older Wavetek 9500 (no "B").  9500/3200 systems.
"I often say that when you can measure what you are speaking about, and express it in numbers, you know something about it; but when you cannot measure it, when you cannot express it in numbers, your knowledge is of a meagre and unsatisfactory kind."
Lord Kelvin (1824-1907)
from lecture to the Institute of Civil Engineers, 3 May 1883

CalLabSolutions

Yes, I recommend you add the test point to both the 9500 and the 9500B.

Tektronix is adding these points to because there is a probe that wants 250k impedance to operate properly. 

Mike.
Michael L. Schwartz
Automation Engineer
Cal Lab Solutions
  Web -  http://www.callabsolutions.com
Phone - 303.317.6670

Hawaii596

Got it.  I just haven't seen one of those yet.  BTW, I wore, for the first time, my CALlAB magazine shirt yesterday.  Very nice.  Black, Short sleeve button down with pocket and CALLAB logo.  I had it in the package and just found it.
"I often say that when you can measure what you are speaking about, and express it in numbers, you know something about it; but when you cannot measure it, when you cannot express it in numbers, your knowledge is of a meagre and unsatisfactory kind."
Lord Kelvin (1824-1907)
from lecture to the Institute of Civil Engineers, 3 May 1883

RFCAL

Mike, You're right! Pulled the cert and there isn't a 250KOhm test point--left myself a note to request it next cal cycle---btw, where is my shirt!

CalLabSolutions

Glad to hear you like the shirt Hawaii.

RF.. You have to write an article for Cal Lab Magazine to get a shirt, hat or pull over.  I wanted them to say "I was published in Cal Lab Magazine."

And yes, there are some labs out there that are running on assumptions they are not validating their software and missing big things like 250k test points. 
Michael L. Schwartz
Automation Engineer
Cal Lab Solutions
  Web -  http://www.callabsolutions.com
Phone - 303.317.6670

Hawaii596

Oops.  Didn't mean to cause a stir.  I wrote an article for Cal Lab Magazine.  I have a framed copy in my office.  If I can think of another one, perhaps I'll get to write another one day.
"I often say that when you can measure what you are speaking about, and express it in numbers, you know something about it; but when you cannot measure it, when you cannot express it in numbers, your knowledge is of a meagre and unsatisfactory kind."
Lord Kelvin (1824-1907)
from lecture to the Institute of Civil Engineers, 3 May 1883

CalLabSolutions

Quote from: Hawaii596 on 03-25-2013 -- 12:11:50
If I can think of another one, perhaps I'll get to write another one day.

Torque wrench calibration is an are I think we need to revist.
Michael L. Schwartz
Automation Engineer
Cal Lab Solutions
  Web -  http://www.callabsolutions.com
Phone - 303.317.6670

USMC kalibrater

Ill write an article on Torque Wrench calibration! PM me with details, Id love the opportunity.  Im a 3rd year mechanical engineering student as well so it may be useful to be published..  8-)
Jason
"Be polite, be professional, but have a plan to kill everybody you meet." -General James Mattis

CalLabSolutions

Michael L. Schwartz
Automation Engineer
Cal Lab Solutions
  Web -  http://www.callabsolutions.com
Phone - 303.317.6670