33K5-4-504-1

Started by MIRCS, 08-02-2004 -- 17:00:06

Previous topic - Next topic

MIRCS

33K5-4-504-1


Table 2

Add 2500S as the primary standard for the calibration. Accuracy ±0.06˚C

3.1 through 3.4 Remain the same.

4.1.1 Connect all Sensors to the TI through the remote cables.

4.1.2 Place the TI Sensors into the Humidity Generator

4.1.3 Set the Humidity Generator for 50% RH and 35˚C and Press RUN, allow generator to stabilize at these temperatures for a minimum of 30 minutes.

4.1.4 Switch to a different sensor and take readings for each remaining sensor.

4.1.5 Repeat steps 4.1.3 and 4.1.4 with the temperature set to 60˚C.

Leave Tables 2 and 3 the same.

Make the current procedure an alternate method.