silv3rstr3

I have never been able to get calibration executive to run these digitizers.  It always times out and gives a calibration error "Max Time exceeded before operation completed".  I also had issues with it not completing the self test yesterday.  I found a driver called NI-Scope and installed it.  Now it passes self test but still times out on the first step.  The 33250A is initialized by cal exec and I'm using short BNC's.  I checked the output tracking on my power splitter and it's good.  Any suggestions?
"They are in front of us, behind us, and we are flanked on both sides by an enemy that out numbers us 29:1. They can't get away from us now!!"
-Chesty Puller

silv3rstr3

In case anyone wanted to know the answer to that question here is what I got from National Instruments.
"I discovered that this error is an issue that we know about and are currently working to fix.  The fix will be coming in a future release of Cal Exec, but I am not positive of when at this point.  We do have a way that you can fix it now: go to ni.com/idnet and install the latest HP 33120A IVI driver on top of Cal Exec.  This is listed in our database as a work around for the issue, so hopefully this should make it work in your version of Cal Exec."