8508A RESISTANCE MEASUREMENT vs. PRT RESISTANCE

Started by wijogr, 06-02-2017 -- 12:01:23

Previous topic - Next topic

wijogr

I'm seeing a difference between using the 4-wire 200 Ohm LoI and 2 kOhm readings vs. the PRT resistance displayed on the right screen of the 8508A in PRT mode. I think I'm missing a setting but can't figure out what it is. The PRT measurement does not use FAST so I have that turned off in the resistance function. Ohms Guard is set for both functions. I=1 mA for all three. Low pass Filter is on. Both resistance ranges have been Zeroed. Any ideas?

Mr M.

Temperature uses 4-wire current reversal resistance measurement, this would be the equivalent of the TRUΩ setting.

bAdbOb

The current changes when you change ranges. I believe 1 mAdc is the usual current used for RTDs.
A.  F.   PMEL 72-81
Malmstrom 73-75, Ramstein 75-77, Edwards 77-81

NC-Cals

The manual has you zero the 200Ω range in LOI (1 mA) and the 2kΩ range in 1 mA. This is outlined on page 3-27 of the user manual. You have to zero the ranges before you switch to PRT. There are also instructions on setting up for 3-wire PRT measurement.

wijogr

Thank you for all the suggestions. I was not using the TRU Ohms setting in resistance.  :-o After running two tests it is working perfectly. Thanks again.